
The LCGC Blog: Thinking Out of the Box on Headspace Sampling for Gas Chromatography
I’m frustrated with static headspace sampling!
I’m frustrated with static headspace sampling!
There have been several instances recently when my groups have been faced with challenges in which static headspace has not been able to deliver an analytical solution or has taken an inordinate amount of time to optimize.
Solid matrices, polar analytes in polar matrices, volatile matrices, very low analyte concentrations, less volatile analytes, and issues around relative response factors (or relative extraction factors) have all been conspiring against us.
We’ve tried a lot of the usual approaches of optimising sample to air volume ratio (b), equilibration time and temperature, agitation ferocity (!), salting out*, co-solvent addition**, and injection volume (actually injection time as we have a loop sampler on our instrument) have been undertaken during our various challenges, and while some approaches have led to better analytical performance, often we have had to use experimental design approaches to deal with the many interactive variables.
Of course I realize that this isn’t the worst thing in the world, and the job of an analytical chemist is that of optimizing experiments with inter-related variables and dealing with challenging problems, however, it seems that if headspace gas chromatography (GC) methods need to be developed, since it is a great deal of effort to get anywhere near decent performance.
This led me to consider other approaches and discussions with other members of the group regarding more “generic” approaches or solutions which offer more comprehensive results.
Often these discussions led to the possibilities of dynamic headspace extraction (sampling) with thermal desorption as the method of sample introductionÂÂ-I have to say I was nervous! To me, dynamic headspace seems even more complicated; extra equipment, all those different adsorbent tubes to choose from, a whole other set of conditions to optimize for the thermal desorption unit, cryogenics or cold trapping-all too much so early after a long holiday break. Perhaps I’m just getting old and grumpy!
However, one or two approaches have grabbed my attention and even begun to give me some hope that there may be more comprehensive techniques available for extraction and introduction of volatile or semi-volatile analytes which can also be quantitative (if required) and may not need large scale optimisation. Let’s see…
Dynamic headspace sampling (DHS) is achieved using a constant flow of purge gas through the headspace of a conventional headspace vial (typically 10- and 20-mL vials as with static headspace sampling, however alternative vial sizes are available if required for optimization of the sample to headspace ratio (b)).
Because the headspace within the vial is constantly purged, the dynamic technique does not rely on a fixed equilibrium within a closed system (such as static headspace) and as analytes are continually removed from the headspace above the sample, this allows further analyte molecules to be released into the headspace. All analytes are trapped onto an adsorbent prior to liberation in the heated desorption unit. Of course, the sample vial may be heated and agitated as with static headspace sampling.
I accept that my comments regarding adsorbent selection as an experimental variable still exist, however adsorbent tubes with multiple packings are available, which takes some work out of this process (see also later). The dry purge stage of the process may also require optimization, however this tends only to be necessary when using aqueous-based matrices and this particular equipment is automated, enabling any experiments to be carried out unattended, even when one is using a design of experiments approach, and hence all of the my working time doesn’t need to be taken up to attend to the equipment! Indeed, with this system, not only can the experimental variable be altered, the type of adsorbents used can also be selected. Therefore the selection of the optimum sorbent will be ready for me following an unattended overnight campaign.
It should be noted that to achieve higher chromatographic efficiency, it is possible to use cryo- or cold-trapping to refocus analytes as they are released from the sorbent trap during thermal desorption. A thermal desorption / cooled inlet combination is shown in Figure 3.
Figure 4 shows the results of a comparative study of headspace techniques for a sample of dry tea [1] which highlights two nice advantages of the basic dynamic headspace technique, it is able to deal with solid samples (1g of dry tea) and, as can be seen, it is more comprehensive and sensitive for this particular application (which is generally true as far as I can tell) compared to static headspace sampling.
An adaptation of any headspace sampling technique is known as the Full Evaporative (Evaporation) Technique (FET). In this technique, a relatively small volume of sample is used (typically <100 mL) with a headspace vial of 10 or 20 mL volume, in which the sample (including the sample matrix) is fully evaporated. This approach can be very useful when the volatile and semi-volatile analytes have higher distribution constants and therefore a preference to remain within the sample matrix or diluent rather than volatilise into the headspace. This is typical when, for example, the sample is aqueous, and the analytes are polar and therefore have a high chemical affinity for the sample matrix or where analytes have a low vapor pressure even under elevated extraction temperatures or increased agitation.
What should be immediately obvious from Figure 6 is the bias of the FET-DHS sample extract towards an increase in compounds at later elution times, which should represent higher boiling or more polar compounds. This correlates with analytes whose distribution constants are higher, and the FET technique obviously presents a better option where analytes are more intractable.
Figure 7 also highlights the further advantage of the FET technique when undertaking quantitative targeted analysis where one can clearly see the sensitivity differences between the two extraction techniques [3].
A further, very interesting technique, is the Multi Volatiles Method (MVM) which represents an excellent way to ensure that all volatile compounds, or indeed all analytes of interest, are identified and quantified, even within highly complex matrices. Using this approach, the full evaporation technique can be used with sequential dynamic headspace extractions under different temperature and flow conditions, and using different thermal desorption trap materials. The idea is to have a comprehensive headspace extraction technique which is capable of extracting analytes with a wide range of differing chemistries and therefore distribution constants. This process is shown in schematic form in Figure 8.
Using the MVM approach, each of the thermal desorption tubes may be analyzed separately to “fractionate” the headspace out, or, when combined with a cryo-inlet, each tube may be desorbed into the inlet and the total output then introduced onto the GC column for a “full profile” type approach. This represents an approach which is much more comprehensive than static headspace GC and Reference 5 outlines various approaches to quantitative analysis using the technique.
The desorption tube materials can be selected to match the analyte chemistries associated with each ‘fraction’ however in most applications I have seen, these tend to be restricted to three different packing varieties.
As the process can be automated, even a full experimental design approach can be taken to the optimization of the various experimental parameters in an overnight campaign.
Figure 9 highlights the comprehensive nature of the data obtained from the FET-MVM approach.
At this point, work continues on a number of these headspace sampling variants in order to identify those that are most suited to our various analytical challenges. While I absolutely concede that the complexity and number of variables, which I was so opposed too in the introduction to this piece, is greater with the DHS-FET and DHS-MVM approaches, the ability to fully automate the development process, the comprehensive nature of the technique, the ability to deal with various matrices and the high sensitivity make them very appealing.
* Just a word on salting out to improve the static headspace recovery of polar analytes from polar matrices. We discovered a very helpful table that describes the relative efficiency of salting out in reference 6 which led us to achieve higher static headspace extraction with a much lower quantity of ammonium sulphate than the sodium chloride that we would have typically chosen.
** We are currently also investigating the use of co-solvents, higher boiling point diluents and other additives that promote analyte partitioning into the headspace, which can be very useful when dealing with non-polar matrices. For more information see references 6 and 7.
References
[1] Comparison of extraction techniques for volatiles in a selection of tea samples, Technical note no. AS188, Anatune, Cambridge UK.
[2] Flavor and Fragrance Analysis of Consumer Products – Dynamic Headspace Compared to Some Traditional Analysis Approaches, AppNote 6/2012, Gerstel GmbH & Co. KG, Eberhard-Gerstel-Platz 1, D-45473 Mülheim an der Ruhr, Germany
[3] Dynamic headspace (DHS) analysis using Full Evaporation Technique (FET) to quantify trace level analytes present in a herbal based liquor, Chromatography Technical Note No AS106, Dan Carrier, Anatune Ltd. Cambrige, UK.
[4] Multi-volatile method for aroma analysis using sequential dynamic headspace sampling with an application to brewed coffee, Nobuo Ochiai, JunTsunokawa, Kikuo Sasamoto, Andreas Hoffmann, Journal of Chromatography A, Volume 1371, 5 December 2014, Pages 65-73
[5] Extension of a dynamic headspace multi-volatile method to milliliter injection volumes with full: sample evaporation: Application to green tea, Nobuo Ochiai, JunTsunokawa, Kikuo Sasamoto, Andreas Hoffmann, Kazunori Okanoya, Kevin MacNamara, Journal of Chromatography A, Volume 1421, 20 November 2014, Pages 103-113
[6] Room temperature ionic liquid as matrix medium for the determination of residual solvents in pharmaceuticals by static headspace gas chromatography, Liu FH, Jiang Y J Chromatogr A 2007 1167: 116-119.
[7] Utilization of a Matrix Effect to Enhance the Sensitivity of Residual Solvents in Static Headspace Gas Chromatography, Zhi Chen et. al. Chen et al., J Chromatogr Sep Tech 2015, 6:6
Tony Taylor is Group Technical Director of Crawford Scientific Group and CHROMacademy. His background is in pharmaceutical R&D and polymer chemistry, but he has spent the past 20 years in training and consulting, working with Crawford Scientific Group clients to ensure they attain the very best analytical science possible. He has trained and consulted with thousands of analytical chemists globally and is passionate about professional development in separation science, developing CHROMacademy as a means to provide high-quality online education to analytical chemists. His current research interests include HPLC column selectivity codification, advanced automated sample preparation, and LC–MS and GC–MS for materials characterization, especially in the field of extractables and leachables analysis.
Newsletter
Join the global community of analytical scientists who trust LCGC for insights on the latest techniques, trends, and expert solutions in chromatography.




