
- LCGC North America-10-01-2019
- Volume 37
- Issue 10
What a GC–MS Tune Report Can Tell You
An essential tool for troubleshooting and tuning GC–MS instruments is the tune report; this article explains the meaning and uses of this report when combined with a tuning compound.
Tuning in gas chromatography-mass spectrometry (GC–MS) involves adjusting several mass spectrometer parameters through the infusion of a tune compound, commonly perfluorotributylamine (PFTBA). The tune report generated is invaluable because it indicates how well the instrument is operating, and is an essential tool when troubleshooting is required. A typical autotune report is shown in Figure 1.
Figure 1: Example autotune report.
There are several parameters that need to be examined in this autotune report:
1. Correct mass assignments. Check that the tune has correctly assigned the masses of peaks 69 (base peak), 219, and 502. The tolerance for unit mass resolution are: 69 (68.8–69.2), 219 (218.8–219.2), 502 (501.8–502.2).
2. The mass peak widths (PW50) should be 0.55 ±0.1.
3. Peaks should be smooth and Gaussian. Do not be concerned with peak shoulders unless they become a major component of the spectral peak. The shoulders should increase in area to reflect the isotopic abundance of C13.
4. Note the electron multiplier (EM) voltage. A clean source with a relatively new EM horn should report a gain voltage of 1400 to 1600. As the lifetime of the dynode increases and the source becomes contaminated, this value may approach values of 3000 V. If the voltage consistently needs to be ramped to 2800–3000 V, it may be time for a source clean or a new dynode.
5. Low background. The presence of a large number of peaks across the spectrum is known as a high background. This can arise from several sources of contamination, such as column bleed, septum bleed, oil contamination, and various other sources.
6. Low water and air. Presence of large peaks at 18, 28, or 32 indicate an air or water leak.
7. The tables below give the relative abundances of each of the ions, as well as isotopic masses, abundances, and ratios. The isotopic mass should always be an M +1 ion, for example, 70, 220, and 503 amu. As PFTBA contains only C13 and N15, the relative isotopic ratios can be easily calculated and should be approximately 69 (1.0), 219 (4.0), and 502 (10.0). Any drift in these ratios can indicate that the spectrometer mass axis has not been correctly calibrated, or that the system resolution is grossly degraded. The following conditions are typical if everything is functioning correctly. The absolute abundance of mass 69 should be ≥200,000 but ≤400,000. For a maximum sensitivity autotune, the relative ratios of each of the peaks should be within the following ranges: 219/69 (20–35%) and 502/69 (0.51%).
8. Proper absolute abundance.
9. The data system will allow the operator to perform an air and water check to quantify the amounts within the spectrometer at any given point in time. This can be a useful tool to check the vacuum system and spectrometer operating temperatures are equilibrated, or to check if a suspected leak has been properly fixed.
All other parameters within the autotune will differ from instrument to instrument, but general trends from subsequent autotune reports should be noted. Any consistent drift in either the positive or negative direction should be noted and then re-autotune, manual tune, or maintenance should be carried out accordingly.
Articles in this issue
about 6 years ago
Ultraviolet Detectors: Perspectives, Principles, and Practicesabout 6 years ago
Sampling: The Ghost in Front of the Laboratory Doorabout 6 years ago
Vol 37 No 10 LCGC North America October 2019 Regular Issue PDFNewsletter
Join the global community of analytical scientists who trust LCGC for insights on the latest techniques, trends, and expert solutions in chromatography.





